BMET Wiki
Advertisement
Electrosurgical unit (Bovie)
Esu


6 Months / 15 Minutes
6 Months / 30 Minutes
6 Months / 60 Minutes
0 Months / 0 Minutes
0 Months / 120 Minutes
High


Electrodes

About[]

application of a high-frequency electric current to biological tissue as a means to cut, coagulate, desiccate, or fulgurate tissue. Its benefits include the ability to make precise cuts with limited blood loss. Electrosurgical units are frequently used during surgical operations helping to prevent blood loss in hospital operating rooms or in outpatient procedures.

In electrosurgical procedures, the tissue is heated by an electric current. Although electrical devices may be used for the cauterization of tissue in some applications, electrosurgery is usually used to refer to a quite different method than electrocautery. The latter uses heat conduction from a probe heated by a direct current (much in the manner of a soldering iron), whereas electrosurgery uses alternating current to directly heat the tissue itself.

Often electrosurgery is mistakenly referred to as diathermy. Unlike Ohmic heating by electric current passing through the conductive tissue in conventional electrosurgery, diathermy means dielectric heating, produced by rotation of molecular dipoles in high frequency alternating electric field. This effect is most widely used in microwave ovens which operate at GHz frequencies.

Electrosurgery is commonly used in dermatological, gynecological, cardiac, plastic, ocular, spine, ENT, orthopedic, urological, neuro- and general surgical procedures.

Electrosurgery is performed using an Electrosurgical Generator (also referred to as Power Supply or Waveform Generator) and a handpiece including one or several electrodes, sometimes referred to as an RF Knife. The apparatus when used for coagulation in surgery is still often referred to informally by surgeons as a "Bovie," after the inventor.

History[]

Development of the first commercial electrosurgical device is credited to Dr. William T. Bovie, who worked on it from 1914 to 1927 while employed at Harvard University[1][2] The first use of an electrosurgical generator in operating room occurred on October 1, 1926. The surgery was performed by Dr. Harvey Williams Cushing.

Clinical application[]

Neural and muscle cells are electrically-excitable, i.e. they can be stimulated by electric current. In human patients such stimulation may cause acute pain, muscle spasms, and even cardiac arrest. Sensitivity of the nerve and muscle cells to electric field is due to the voltage-gated ion channels present in their cell membranes. Stimulation threshold does not vary much at low frequencies (so called rheobase - constant level). However, the threshold starts increasing with decreasing duration of a pulse (or a cycle) when it drops below a characteristic minimum (so called chronaxie). Typically, chronaxie of neural cells is in the range of 0.1 - 10 ms, so the sensitivity to electrical stimulation (inverse of the stimulation threshold) decreases with increasing frequency in the kHz range and above. (Note that frequency of the alternating electric current is an inverse of the duration of a single cycle). To minimize the effects of muscle and neural stimulation, electrosurgical equipment typically operates in the radio frequency (RF) range of 100 kHz to 5 MHz.

Operation at higher frequencies also helps minimizing the amount of hydrogen and oxygen generated by electrolysis of water. This is especially important consideration for applications in liquid medium in closed compartments, where generation of gas bubbles may interfere with the procedure. For example, bubbles produced during an operation inside an eye may obscure a field of view.

Common electrode configurations[]

There are several commonly used electrode configurations or circuit topologies:

In bipolar configuration the voltage is applied to the patient using a pair of similarly-sized electrodes. For example, special forceps, with one tine connected to one pole of the AC generator and the other tine connected to the other pole of the generator. When a piece of tissue is held by the forceps, a high frequency electric current flows from one to the other forceps tine, heating the intervening tissue.

In monopolar configuration the patient lies on top of the return electrode, a relatively large metal plate or a flexible metalized plastic pad which is connected to the return electrode of the AC source. The surgeon uses a pointed probe to make contact with the tissue. The electric current flows from the probe tip, through the body to the return electrode, and then back to the electrosurgical generator. Since electric current spreads from the pointed electrode as it enters the body the current density is rapidly (quadratically) decreasing with distance from the electrode. Since the rate of heating is proportional to the square of current density, the heating occurs in a very localized region, only near the probe tip. On an extremity such as a finger, there is limited cross-sectional area for the return current to spread across, which might result in higher current density and some heating throughout the volume of the extremity.


Relatively low-powered high frequency electrosurgery can be performed on conscious outpatients with no return electrode at all [3]. Operating with no return electrode is possible, because at the very high frequencies and low currents, the self-capacitance of the patient's body (which is between the patient's body and the machine's return potential) is large enough to allow the resulting displacement current to act as a return path. One example of such a machine is called a hyfrecator.

An accidental additional return path through an earth-ground provides a danger of a burn at a site far away from the probe electrode, and for this reason single-electrode devices are used only on conscious patients who would be aware of such complications, and only on carefully insulated tables.

Electrosurgical modalities[]

In cutting mode electrode touches the tissue, and sufficiently high power density is applied to vaporize its water content. Since water vapor is not conductive under normal cirumstances, electric current cannot flow through the vapor layer. Energy delivery beyond the vaporization threshold can continue if sufficiently high voltage is applied (> +/-200 V) to ionize vapor and convert it into a conductive plasma. Vapor and fragments of the overheated tissue are ejected, forming a crater. Electrode surfaces intended to be used for cutting often feature a finer wire or wire loop, as opposed to a more flat blade with a rounded surface.

Coagulation is performed using waveforms with lower average power, generating heat insufficient for explosive vaporization, but producing a thermal coagulum instead.

Electrosurgical desiccation occurs when the electrode touches the tissue open to air, and the amount of generated heat is lower than that required for cutting. The tissue surface and some of the tissue more deep to the probe dries out and forms a coagulum (a dry patch of dead tissue). This technique may be used for treating nodules under the skin where minimal damage to the skin surface is desired.

In fulguration mode, the electrode is held away from the tissue, so that when the air gap between the electrode and the tissue is ionized, an electric arc discharge develops. In this approach the burning to the tissue is more superficial, because the current is spread over the tissue area larger than the tip of electrode. [4] Under these conditions, superficial skin charring or carbonization is seen over a wider area than when operating in contact with the probe, and this technique is therefore used for very superficial or protrusive lesions such as skin tags. Ionization of an air gap requires voltage in the kV range.

Besides the thermal effects in tissue, electric field can produce pores in the cellular membranes - a phenomenon called electroporation. This effect may affect cells beyond the range of thermal damage.

Electrosurgical waveforms[]

Different waveforms can be used for different electrosurgical procedures. For cutting, a continuous single frequency sine wave is often employed. Rapid tissue heating leads to explosive vaporization of interstitial fluid. If the voltage is sufficiently high (> 400 V peak-to-peak)the vapor sheath is ionized, forming conductive plasma. Electric current continues to flow from the metal electrode through the ionized gas into the tissue. Rapid overheating of tissue results in its vaporization, fragmentation and ejection of fragments, allowing for tissue cutting in applications of a continuous wave the heat diffusion typically leads to formation of a significant thermal damage zone at the edges of the lesion. Open circuit voltage in electrosurgical waveforms is typically in the range of 300 - 10,000 V peak-to-peak.

Higher precision can be achieved with pulsed waveforms using bursts of several tens of microseconds in duration the tissue can be cut, while the size of the heat diffusion zone does not exceed the cellular scale. Heat accumulation during repetitive application of bursts can also be avoided if sufficient delay is provided between the bursts, allowing the tissue to cool down.

The proportion of ON time to OFF time can be varied to allow control of the heating rate. A related parameter, duty cycle, is defined as the ratio of the ON time to the period (the time of a single ON-OFF cycle). In the terminology of electrical engineering, this process of altering an amplitude of a periodic waveform is called modulation.

For coagulation, the average power is typically reduced below the threshold of cutting. Typically, sine wave is turned ON and OFF in a rapid succession. The overall effect is a slower heating process, which causes tissue to coagulate. In simple coagulation/cutting mode machines, the lower duty cycle typical of coagulation mode is usually heard by the ear as a lower frequency and a rougher tone, than the higher frequency tone typical of cutting mode with the same equipment.

Many modern electrosurgical generators provide sophisticated waveforms with power adjusted in real time, based on changes of the tissue impedance.

Prevention of unintended burns in patients[]

For high power surgical uses during anesthesia the monopolar modality relies on a good electrical contact between a large area of the body (typically at least the entire back of the patient) and the return electrode. If contact with the return pad is insufficient, severe burns (3rd degree) can occur in areas of poor contact with the return pad, or with metal objects in contact with Earth-ground serving as an unintended (capacitative) return path. It is preferrable to use patient return electrodes that are "split" and have the capacity to be used with a compatible Electrosurgical generator that will monitor the patient/pad interface ensuring that the interface is adequately performing, to significantly decrease the liklihood of a pad site/unintended site burn.

To prevent unintended burns, the skin is cleaned and a conductive gel is used to enhance contact. Proper electrical grounding practices must be followed in the electrical wiring of the building. It is also recommended to use a newer electrosurgical unit that includes alarms for ground circuit interruption. Grounding pads should always have full contact with the skin and be placed on the same side of the body and close to the body part where the procedure is occurring.

If there is any metal in the body of the patient, the grounding pad is placed on the opposite side of the body from the metal and be placed between the metal and the operation site. This prevents current from passing selectively through metal on the way to ground. For example, for a patient who has had a right sided hip replacement who is scheduled for surgery, the grounding pad is placed on the left side of the body on the lateral side of the lower abdomen, which places the grounding pad between the location of the metal and the surgical site and on the opposite side from the metal. If there is metal on both sides of the body, the grounding pad is placed between the metal and the procedure site when possible. Common grounding pad locations include lateral portions of the outer thighs, abdomen, back, or shoulder blades. [1] The use of the bipolar option does not require the placement of a grounding pad because the electrical current does not flow through the bulk of the body. Electrosurgery should only be performed by a physician who has received specific training in this field and who is familiar with the techniques used to prevent burns.

Manufacturers[]

Models[]

  • Force 2, Valleylab
  • Force FX, Valleylab
  • System 5000
  • Sabre 180

Links[]

Covidien Energy-Based Professional Education


Second Source Parts[]

  • Parts Source LLC

Second Source Service[]

  • BiTech Medical
  • 2484 Briarcliff Rd. Ne
  • Atlanta, Ga
  • 888-470-3720

References[]

  1. 1.0 1.1 Understanding Electrosurgery, Genard McCauley, Aaron Medical.
  2. Sheldon V. Pollack FRCPC, Alastair Carruthers FRCPC, Roy C. Grekin MD (2000) "The History of Electrosurgery", Dermatologic Surgery 26 (10), 904–908.
  3. see page 6
  4. Electrosurgery for the Skin. Barry L. Hainer M.D., Richard B. Usatine, M.D., American Family Physician (Journal of the American Academy of Family Physicians), 2002 Oct 1;66(7):1259-66. See illustration.

Links[]

See also[]

ESU Analyzer

Video[]

Bovie

Advertisement